... As part of a quality control procedure prior to a clinical trial we measured the delivered [Na]d using a flame photometry assay. Dialysate was sampled cyclically over a 6-month period from 71 dialysis machines. 32/71 [35.7%] machines were delivering a mean [Na]d >144mmol/L. The machines were recalibrated with fortnightly assessment of delivered [Na]d until they were at target 140mmol/l within the tolerances of the machine [2 mmol/L].Post calibration there was a significant decrease in mean [Na]d, 143.52.1 v 141.5 1.4 mmol/L [p<0.0001] and only 6.7% dialysis machines continued to deliver [Na]d >144mmol/L ...