Hemolysis during dialysis can be attributed to other factors such as hypotonic dialysate, choloramine levels, semioccluded/ defective tubing/dialyzer (
http://www.nature.com/ki/journal/v57/n4/full/4491507a.html ) rather than pump pressure, but here are the numbers:
"Therefore, a conservative approach to quality care for hemodialysis patients would be to keep the
prepump arterial pressure less negative than -260 mmHg. This would ensure an adequate delivery of blood to the dialyzer and decrease the potential of hemolysis due to shear stress or negative pressure.
Risks associated with excessive negative pressure as monitored by the prepump arterial pressure monitor device are:
1. hemolysis due to excessive negative pressure and shearing of the blood cells;
.............
Risks as a result of excessive
(venous) pressure, defined in the ANNA Core Curriculum as
greater than 400 mmHg (Keen et al., 1995, p. 225), at the post-pump arterial pressure monitor are:
1. hemolysis due to excessive pressure and shearing of the blood cells; and..... "