I Hate Dialysis Message Board
Welcome, Guest. Please login or register.
November 24, 2024, 08:10:06 AM

Login with username, password and session length
Search:     Advanced search
532606 Posts in 33561 Topics by 12678 Members
Latest Member: astrobridge
* Home Help Search Login Register
+  I Hate Dialysis Message Board
|-+  Dialysis Discussion
| |-+  Dialysis: General Discussion
| | |-+  venous pressure.....
0 Members and 1 Guest are viewing this topic. « previous next »
Pages: [1] Go Down Print
Author Topic: venous pressure.....  (Read 2778 times)
tyefly
Elite Member
*****
Offline Offline

Gender: Female
Posts: 2016


This will be me...... Next spring.... I earned it.

« on: October 31, 2010, 03:26:08 PM »

Ok   for the third day in a row..... I had to quit my treatment because of very high venous pressure..... what is going on...... 400 or better.....   even when I slow the speed down the pressure reduce alittle but back up....  having alarm 30.... today it happened with in 20 min of starting treatment...yesterday I had to stop in a hour.... I called nxstage and they suggested that I put a syring on a pull out blood in my venous.... no problem....  came out easy and went in easy......   I even increased my heparin use today  up from 3000 to 4000..... 
  I am worried.....   I have had a good treatment in three days....    is there something wrong with the machine...or my fistula.....  do I need more heparin.....     do I need a angiogram.....   am I going to get one of the neck hook ups.....( I have never had one of those )     

   Ok   Yes I am a little scared......  after all     this is my life line......   I hate the unknown.....
Logged

IgA Nephropathy   April 2009
CKD    May 2009
AV Fistula  June 2009
In-Center Dialysis   Sept 2009
Nxstage    Feb 2010
Extended Nxstage March 2011

Transplant Sept 2, 2011

  Hello from the Oregon Coast.....

I am learning to live close to the lives of my friends without ever seeing them. No miles of any measurement can separate your soul from mine.
- John Muir

The clearest way into the Universe is through a forest wilderness.
- John Muir
BigSky
Elite Member
*****
Offline Offline

Gender: Male
Posts: 2380


« Reply #1 on: October 31, 2010, 04:13:20 PM »

Could be a stenosis starting in your fistula with pressures that high.

Also part of the problem can be that Nx Stage seems to have a problem in measuring pressure IMO, but again with pressures that high I would be weary of stenosis


On Cobe and Phoenix machines my venous is 220-230 running at a blood flow of 450 and blood pump speed set at 500.

Nx Stage venous is 250 with a blood speed of 400. 



I might add,  you most likely will not need a neck or chest cath if they do a fistulagram unless they damage it.  Usually they can do a fistulagram in an hour or so and then one can dialyze immediately afterword.
« Last Edit: October 31, 2010, 04:35:35 PM by BigSky » Logged
Rerun
Member for Life
******
Offline Offline

Gender: Female
Posts: 12242


Going through life tied to a chair!

« Reply #2 on: October 31, 2010, 04:25:41 PM »

Gosh, if you called NxStage didn't they have any answers??

              :waiting;
Logged

greg10
Full Member
***
Offline Offline

Gender: Male
Posts: 469


« Reply #3 on: October 31, 2010, 05:32:05 PM »

Gosh, if you called NxStage didn't they have any answers??
              :waiting;
NxStage will give you answers to the workings of the dialysis cycler and dialysate.  They will not be able to answer too much about the state of the patient and their access.

Unfortunately, an access that is not working well will cause a venous side high pressure.  This could be due to clots or cannulation.  Have you called your local dialysis center that trained you on NxStage and ask then about your potential access problem?  I find that they are usually the ones with the answers that can help to get your access working properly. 
Logged

Newbie caretaker, so I may not know what I am talking about :)
Caretaker for my elderly father who has his first and current graft in March, 2010.
Previously in-center hemodialysis in national chain, now doing NxStage home dialysis training.
End of September 2010: after twelve days of training, we were asked to start dialyzing on our own at home, reluctantly, we agreed.
If you are on HD, did you know that Rapid fluid removal (UF = ultrafiltration) during dialysis is associated with cardiovascular morbidity?  http://ihatedialysis.com/forum/index.php?topic=20596
We follow a modified version: UF limit = (weight in kg)  *  10 ml/kg/hr * (130 - age)/100

How do you know you are getting sufficient hemodialysis?  Know your HDP!  Scribner, B. H. and D. G. Oreopoulos (2002). "The Hemodialysis Product (HDP): A Better Index of Dialysis Adequacy than Kt/V." Dialysis & Transplantation 31(1).   http://www.therenalnetwork.org/qi/resources/HDP.pdf
tyefly
Elite Member
*****
Offline Offline

Gender: Female
Posts: 2016


This will be me...... Next spring.... I earned it.

« Reply #4 on: October 31, 2010, 05:33:49 PM »

Yes called nxstage.....   they had me unhook and place a syring and pull blood.... I did not have any problems with that.... hooked up again ....and  U 400 to 500.      I had those pressures even at Blood speed 250....  Nxstage told me to quit  and contact my center......   so thats what I did.... 

     I will call them in the morning......
Logged

IgA Nephropathy   April 2009
CKD    May 2009
AV Fistula  June 2009
In-Center Dialysis   Sept 2009
Nxstage    Feb 2010
Extended Nxstage March 2011

Transplant Sept 2, 2011

  Hello from the Oregon Coast.....

I am learning to live close to the lives of my friends without ever seeing them. No miles of any measurement can separate your soul from mine.
- John Muir

The clearest way into the Universe is through a forest wilderness.
- John Muir
tyefly
Elite Member
*****
Offline Offline

Gender: Female
Posts: 2016


This will be me...... Next spring.... I earned it.

« Reply #5 on: November 02, 2010, 09:52:14 AM »

So I am going in for a angiogram...fistualgram..what ever its called.....    My center told me that its probably the cartridge that I am using....  which I am using the 171B.....   in prep for extended ( when ever they can find the time to get the heparin pump)....  The clinic told me that they have never used this cartridge with anyone and so they think that it may be the problem ... they told me to switch back to the reg cartridge......  and that since that cartridge causes high venous pressure that I might not be able to do the extended..... 

Hey   any of you that do extended have problems with their cartridges.....    ( I personally dont think this is the problems... but I am not the expert....  but is my clinic....???)

well lets see what the fistulagram test show..... go from there.....   I HATE  Dialysis..........
Logged

IgA Nephropathy   April 2009
CKD    May 2009
AV Fistula  June 2009
In-Center Dialysis   Sept 2009
Nxstage    Feb 2010
Extended Nxstage March 2011

Transplant Sept 2, 2011

  Hello from the Oregon Coast.....

I am learning to live close to the lives of my friends without ever seeing them. No miles of any measurement can separate your soul from mine.
- John Muir

The clearest way into the Universe is through a forest wilderness.
- John Muir
greg10
Full Member
***
Offline Offline

Gender: Male
Posts: 469


« Reply #6 on: November 02, 2010, 10:42:57 AM »

So I am going in for a angiogram...fistualgram..what ever its called.....    My center told me that its probably the cartridge that I am using....  which I am using the 171B..... ..........
You don't have an extra 170B around that you can try for one treatment?  I doubt that is the reason because the accompanying 170B document describes the 171B as well, the only difference I could see was the extra T line for the syringe pump.

We have not tried the 171B.
Logged

Newbie caretaker, so I may not know what I am talking about :)
Caretaker for my elderly father who has his first and current graft in March, 2010.
Previously in-center hemodialysis in national chain, now doing NxStage home dialysis training.
End of September 2010: after twelve days of training, we were asked to start dialyzing on our own at home, reluctantly, we agreed.
If you are on HD, did you know that Rapid fluid removal (UF = ultrafiltration) during dialysis is associated with cardiovascular morbidity?  http://ihatedialysis.com/forum/index.php?topic=20596
We follow a modified version: UF limit = (weight in kg)  *  10 ml/kg/hr * (130 - age)/100

How do you know you are getting sufficient hemodialysis?  Know your HDP!  Scribner, B. H. and D. G. Oreopoulos (2002). "The Hemodialysis Product (HDP): A Better Index of Dialysis Adequacy than Kt/V." Dialysis & Transplantation 31(1).   http://www.therenalnetwork.org/qi/resources/HDP.pdf
tyefly
Elite Member
*****
Offline Offline

Gender: Female
Posts: 2016


This will be me...... Next spring.... I earned it.

« Reply #7 on: November 02, 2010, 09:59:11 PM »

Had the fistulagram.....   they infiltrated me    and so they couldnt do very much testing...... Gee I havnt even sone that to my self yet.....  Doc said   that I sould be just fine..... just make sure when I dialysis next time make sure my arm does not move and keep it in a downward motion ...  just make sure I dont move it at all during the run......         oh  Yea   ...........thats what I am going to do .....sure I am ...... 
 I did run successfully with the 170 cartridge....going to try the 171 again .....   I cant believe its the cartridge...
Logged

IgA Nephropathy   April 2009
CKD    May 2009
AV Fistula  June 2009
In-Center Dialysis   Sept 2009
Nxstage    Feb 2010
Extended Nxstage March 2011

Transplant Sept 2, 2011

  Hello from the Oregon Coast.....

I am learning to live close to the lives of my friends without ever seeing them. No miles of any measurement can separate your soul from mine.
- John Muir

The clearest way into the Universe is through a forest wilderness.
- John Muir
Bruno
Full Member
***
Offline Offline

Gender: Male
Posts: 377


TOFF (typical old Fart)

« Reply #8 on: November 11, 2010, 02:13:32 AM »

Maybe your blood pump speeds are too high.
Logged
Pages: [1] Go Up Print 
« previous next »
 

Powered by MySQL Powered by PHP SMF 2.0.17 | SMF © 2019, Simple Machines | Terms and Policies Valid XHTML 1.0! Valid CSS!